A sharper X-ray image with a dose trade-off

A Level extension · Original GioPhysics question with a detailed, mark-by-mark answer guide.

Demand
discriminating
Marks
6
Topics
1
Answer
Complete
Explore the question data

Measured X-ray attenuation

Supplied question data. Read the labelled axes and use the observations to test the physical model.

Vertical axis (y): I/I₀

Measured X-ray attenuationSupplied question data. Read the labelled axes and use the observations to test the physical model. Horizontal axis: Thickness / cm. Vertical axis: I/I₀. Exact values are available in the data table below.00.250.50.75100.751.52.253xy

Horizontal axis (x): Thickness / cm

Reading 1Thickness / cm: 0I/I₀: 1
Read the exact data as a table
Measured X-ray attenuation — supplied values
Thickness / cmI/I₀
01
10.5
20.25
30.125

These are the supplied values, not a worked answer. Lines join the supplied values; they are not a fitted model.

extended response6 marks

An X-ray beam passes through a uniform absorber. Measured transmitted intensity fractions are supplied. Assume exponential attenuation I/I0 = exp(−μx), ignore scatter, and use h = 6.63×10⁻³⁴ J s, c = 3.0×10⁸ m s⁻¹ and electron charge e = 1.60×10⁻¹⁹ C. The tube voltage is 60 kV. Supplied model: For the shortest X-ray wavelength use eV = hc/λmin. Supplied data (Thickness / cm; I/I₀): (0, 1), (1, 0.5), (2, 0.25), (3, 0.125).

  1. (a)

    Determine attenuation coefficient, transmission through 4 cm and the shortest possible tube-emission wavelength.

    3 marks
  2. (b)

    Explain the image contrast and evaluate raising beam intensity to improve a weak image.

    3 marks
Ready to self-mark?Reveal the detailed answer guide
Answer overview(a) μ≈69.3 m⁻¹; transmission at 4 cm = 1/16 = 6.25%; λmin≈2.07×10⁻¹¹ m. (b) Different attenuation gives contrast; more incident photons can reduce relative counting noise but also increases exposure.
01

(a)

3 marks

Determine attenuation coefficient, transmission through 4 cm and the shortest possible tube-emission wavelength.

How to approach it

Build a chain of claim, evidence and physics reasoning. Address more than one relevant factor, identify a limitation or assumption, and finish with a conclusion that is conditional on the evidence rather than absolute.

  1. 1

    The half-value thickness is 0.01 m, so μ = ln2/0.01 ≈69.3 m⁻¹. Four such thicknesses transmit (1/2)⁴ = 1/16.

  2. 2

    The greatest photon energy is the full electron energy eV = (1.60×10⁻¹⁹)(60000) = 9.60×10⁻¹⁵ J.

  3. 3

    Using E = hc/λ gives λmin = hc/(eV) ≈2.07×10⁻¹¹ m. Most emitted photons need not have this maximum energy.

02

(b)

3 marks

Explain the image contrast and evaluate raising beam intensity to improve a weak image.

How to approach it

Build a chain of claim, evidence and physics reasoning. Address more than one relevant factor, identify a limitation or assumption, and finish with a conclusion that is conditional on the evidence rather than absolute.

  1. 1

    Different tissue thicknesses and attenuation coefficients give different transmitted intensities at the detector, creating contrast.

  2. 2

    If detected photons are independently counted, relative random counting uncertainty scales approximately as 1/√N; increasing N can improve statistical image quality.

  3. 3

    X-rays are ionising, so increased photon fluence can increase patient dose. The exposure should be justified by the diagnostic task, with collimation, shielding and detector efficiency considered; ultrasound may answer some questions without ionising radiation.

Complete marking guidance

Use these worked steps alongside the part-by-part answers. They explain the method and mark allocation; they do not add extra marks.

  1. The half-value thickness is 0.01 m, so μ = ln2/0.01 ≈69.3 m⁻¹. Four such thicknesses transmit (1/2)⁴ = 1/16.
  2. The greatest photon energy is the full electron energy eV = (1.60×10⁻¹⁹)(60000) = 9.60×10⁻¹⁵ J.
  3. Using E = hc/λ gives λmin = hc/(eV) ≈2.07×10⁻¹¹ m. Most emitted photons need not have this maximum energy.
  4. Different tissue thicknesses and attenuation coefficients give different transmitted intensities at the detector, creating contrast.
  5. If detected photons are independently counted, relative random counting uncertainty scales approximately as 1/√N; increasing N can improve statistical image quality.
  6. X-rays are ionising, so increased photon fluence can increase patient dose. The exposure should be justified by the diagnostic task, with collimation, shielding and detector efficiency considered; ultrasound may answer some questions without ionising radiation.