An ultrasound image with two hidden boundaries

Physics in society · Original GioPhysics question with a detailed, mark-by-mark answer guide.

Demand
discriminating
Marks
6
Topics
1
Answer
Complete
Explore the question data

Echo timing evidence

Supplied question data. Read the labelled axes and use the observations to test the physical model.

Vertical axis (y): Relative echo amplitude

Echo timing evidenceSupplied question data. Read the labelled axes and use the observations to test the physical model. Horizontal axis: Time after transmission / μs. Vertical axis: Relative echo amplitude. Exact values are available in the data table below.00.250.50.751018.7537.556.2575xy

Horizontal axis (x): Time after transmission / μs

Reading 1Time after transmission / μs: 0Relative echo amplitude: 1
Read the exact data as a table
Echo timing evidence — supplied values
Time after transmission / μsRelative echo amplitude
01
50
350
400.6
450
650
700.3
750

These are the supplied values, not a worked answer. Lines join the supplied values; they are not a fitted model.

extended response6 marks

Optional enrichment: this challenge may extend beyond the usual syllabus depth and assumes confidence with algebra and graph interpretation. A pulse-echo scanner detects reflected pulses at 40 μs and 70 μs after transmission into tissue. Sound speed is 1500 m s⁻¹. A 5.0 MHz pulse lasts four cycles. At one boundary acoustic impedances are Z1 = 1.5×10⁶ and Z2 = 6.0×10⁶ kg m⁻² s⁻¹. Use reflected intensity fraction R = [(Z2−Z1)/(Z2+Z1)]². Supplied data (Time after transmission / μs; Relative echo amplitude): (0, 1), (5, 0), (35, 0), (40, 0.6), (45, 0), (65, 0), (70, 0.3), (75, 0).

  1. (a)

    Find both boundary depths and their separation, assuming the same wave speed throughout the path.

    3 marks
  2. (b)

    Estimate axial pulse resolution and the reflection fraction. Explain the role of coupling gel and one access or performance trade-off.

    3 marks
Ready to self-mark?Reveal the detailed answer guide
Answer overview(a) Depths = 30 mm and 52.5 mm; separation = 22.5 mm. (b) Pulse duration = 0.80 μs; axial resolution≈0.60 mm; reflected intensity fraction = 0.36.
01

(a)

3 marks

Find both boundary depths and their separation, assuming the same wave speed throughout the path.

How to approach it

Build a chain of claim, evidence and physics reasoning. Address more than one relevant factor, identify a limitation or assumption, and finish with a conclusion that is conditional on the evidence rather than absolute.

  1. 1

    Each pulse travels to a boundary and back, so depth is d = vt/2.

  2. 2

    At 40 μs, d = 1500(40×10⁻⁶)/2 = 0.030 m. At 70 μs, d = 0.0525 m.

  3. 3

    Their separation is 0.0525−0.030 = 0.0225 m. Different tissue speeds or refraction would make this uniform-speed interpretation approximate.

02

(b)

3 marks

Estimate axial pulse resolution and the reflection fraction. Explain the role of coupling gel and one access or performance trade-off.

How to approach it

Build a chain of claim, evidence and physics reasoning. Address more than one relevant factor, identify a limitation or assumption, and finish with a conclusion that is conditional on the evidence rather than absolute.

  1. 1

    Four cycles at 5.0 MHz last 4/(5.0×10⁶) = 0.80 μs. The spatial pulse length is 1500(0.80×10⁻⁶) = 1.2 mm; two interfaces need about half this separation, 0.60 mm, to resolve distinct echoes.

  2. 2

    R = [(6.0−1.5)/(6.0+1.5)]² = (0.60)² = 0.36. Amplitude ratios and intensity ratios must not be confused.

  3. 3

    Gel removes an air gap that otherwise strongly reflects sound. Ultrasound avoids ionising radiation and can be portable, but penetration, operator skill and acoustic access limit some examinations; higher frequency can improve resolution while increasing attenuation.

Complete marking guidance

Use these worked steps alongside the part-by-part answers. They explain the method and mark allocation; they do not add extra marks.

  1. Each pulse travels to a boundary and back, so depth is d = vt/2.
  2. At 40 μs, d = 1500(40×10⁻⁶)/2 = 0.030 m. At 70 μs, d = 0.0525 m.
  3. Their separation is 0.0525−0.030 = 0.0225 m. Different tissue speeds or refraction would make this uniform-speed interpretation approximate.
  4. Four cycles at 5.0 MHz last 4/(5.0×10⁶) = 0.80 μs. The spatial pulse length is 1500(0.80×10⁻⁶) = 1.2 mm; two interfaces need about half this separation, 0.60 mm, to resolve distinct echoes.
  5. R = [(6.0−1.5)/(6.0+1.5)]² = (0.60)² = 0.36. Amplitude ratios and intensity ratios must not be confused.
  6. Gel removes an air gap that otherwise strongly reflects sound. Ultrasound avoids ionising radiation and can be portable, but penetration, operator skill and acoustic access limit some examinations; higher frequency can improve resolution while increasing attenuation.